The systems also contain 0.2 micron point of use filters which can mask the level of microbiological contamination in the system. While it is recognized that endotoxins are the primary concern in such a system, a filter will reduce microbiological contamination, but not necessarily endotoxin contamination. If filters are used in a water system there should be a stated purpose for the filter, i.e., particulate removal or microbial reduction, and an SOP stating the frequency with which the filter is to be changed which is based on data generated during the validation of the system.
选自FDA 《guide to inspection of highly purified water systems》 |
|